21st Mar 2016, 8:10 pm | #1 | |
Calibrating a Thurlby PL320
Now that I have recapped both PL320 PSU units I decided to check the calibration. I have the older units in the brown case and they have likely not been checked for a long time although they did not seem very far out - just a few millivolts on the DC voltage.
I first used a voltage reference to calibrate the voltage reading on my DMM and then used this to check the PSU following the Thurlby instructions. I believe I have identified the correct service manual version, but I am unsure about the last step. This reads as follows: Quote:
21st Mar 2016, 8:20 pm | #2 |
Re: Calibrating a Thurlby PL320
Don't worry about
This is just a fine-tweak of the offset voltage of the current regulator section. In theory, if it's wildly wrong, then the preset current (DC output switch off) wouldn't match the actual current limit (when DC output switch is on). Whenever I've made PSUs of this topology, I've not bothered to include the adjustment. I did one for a course we run at work, and over the last decade or so, we've probably built between 50 and 100 of them. Very, very occasionally we find one that differs by a couple of counts, and to fix it, I just swap out the op-amp! But the spec is quite loose anyway (+/-2 counts). If you're interested, a (very) old version of the handout is here: https://dl.dropboxusercontent.com/u/...-%20Mk%202.pdf and the schematic is here: https://dl.dropboxusercontent.com/u/9900549/PSU.gif Lining up those PL320s is a nightmare - the presets have too much adjustment range and are far too "touchy". Classic design error. Good luck |
22nd Mar 2016, 12:06 pm | #3 |
Yes, although one or two of those presets were quite straightforward to set, a number of them were rather 'touchy' and so tricky to set accurately. Thanks for your insight and I will not worry about it too much. I have also downloaded your handout and schematic and will have a look at them later.