Before refitting Board 3, decided to monitor the 2 regulator neons.
V301 voltage is 84.8V and V302 is 87.4V (V302 a little higher than spec). The current drain on V301 at min brightness is 0.25mA and at max brightness increases to 1.11mA, with normal viewing brightness the current is around 0.27mA Disconnecting A3U reduces the current to a very low value, but toggling +ve/-ve The current drain on V302 (the lower not visible trace) at min brightness is 0.25mA and at max brightness increases to around 0.9mA Disconnecting A3U reduces the current (once again) to a very low value but toggling +ve/-ve, I disconnected A3U by mistake , meant to have disconnected A3L. David |