Default AVO 8 fault

I have an Avo 8 with a rather baffling fault. I can be using it one day to do various voltage and resistance measurements with no problems. I then leave it undisturbed on the bench for a few days, come back to it and set it up to do a resistance or continuity measurement and it won’t work. If I set it to do a voltage measurement it will work, then it will also work on the resistance ranges. I can then use it all day with no problems.. One might think the meter movement is sticky or there is an intermittent connection somewhere but during the fault condition the movement quite free to move, tapping the case in various places doesn’t cure the fault. I’ve opened the instrument up and had a good look round but can’t see anything suspicious.. It may be temperature related as my electronics workshop is in my loft which suffers wide temperature changes.

Has any other AVO owner experienced similar problems with their instrument?

Michael D
Electrone is offline  
Old 8th Sep 2013, 5:28 pm   #2
threeseven
Octode
 
threeseven's Avatar
 
Join Date: May 2010
Location: Worcestershire, UK.
Posts: 1,880
Default Re: AVO 8 fault

This is likely to be one of the following, intermittent test lead or leads, intermittent cut contacts or a loose contact on the meter movement.
Next time it happens and before operating the switches, try different leads. If you are confident it isn't that, operate the cutout a few times to clean the contacts. After that it's inside to check for faulty switches or meter connections.
__________________
BVWS member

Last edited by threeseven; 8th Sep 2013 at 5:36 pm.
threeseven is offline  
Old 8th Sep 2013, 6:10 pm   #3
pmmunro
Octode
 
Join Date: Jul 2004
Location: Dundee, UK.
Posts: 1,812
Default Re: AVO 8 fault

Michael,

What mark of the Model 8 is your meter? As the internal construction varies completely, this has a significant bearing on likely faults. Threeseven's advice is good, especially on the marks 1 to 4. You could also try pressing firmly on the "reverse moving coil" button a couple of times.

PMM.
pmmunro is offline  
Old 8th Sep 2013, 6:14 pm   #4
Phil G4SPZ
Dekatron
 
Phil G4SPZ's Avatar
 
Join Date: Apr 2005
Location: Bewdley, Worcestershire, UK.
Posts: 4,748
Default Re: AVO 8 fault

I have experienced several faults of this nature with Model 8s, and in most cases it has been caused by poor or dirty leaf switch contacts. These switches are operated by the cams behind the rotary switches and they can be difficult to get at.

Resist the temptation to spray switch cleaner all over them. I bend a pipe cleaner in the middle, dip it in meths then pull the pipe cleaner between the contact points. Look closely at the contacts and rotate the switch until they open so that the pipe cleaner will pass through. Don't use abrasive cleaners or emery cloth. The same technique works for the cut-out contacts, but these have a self-cleaning wiping action and unless the cut-out has operated to clear a heavy overload they should be OK.
__________________
Phil

Optimist [n]: One who is not in possession of the full facts
Phil G4SPZ is offline  
Old 8th Sep 2013, 6:23 pm   #5
threeseven
Octode
 
threeseven's Avatar
 
Join Date: May 2010
Location: Worcestershire, UK.
Posts: 1,880
Default Re: AVO 8 fault

I was indeed assuming it was a pre Mk5 model.
__________________
BVWS member
threeseven is offline  
Old 9th Sep 2013, 10:03 pm   #6
MrElectronicman
Heptode
 
MrElectronicman's Avatar
 
Join Date: Feb 2011
Location: Washington DC, USA
Posts: 619
Default Re: AVO 8 fault

Phil is right, I have and continue to have the same problem with my AVO 8 MK11. The leaf switches are definitely the source of the problem. I have cleaned mine as Phil says and it works for a short time and then after a period of no-use it becomes intermittent again.

The real fix I believe is to dismantle the individual leaf switch assembly and burnish the contacts and then reassemble and adjust the leaf contact pressure. BUT! that is quite a task and not for the faint-hearted.
__________________
David
MrElectronicman is offline  
Old 9th Sep 2013, 10:25 pm   #7
Electrone
Retired Dormant Member
 
Electrone's Avatar
 
Join Date: Aug 2012
Location: Cromer, Norfolk, UK.
Posts: 100
Default Re: AVO 8 fault

My AVO 8 is a MkII. The test leads are definitely OK. At first I thought the cutout was perhaps making intermittent contact and shorted it out withh a wire link. Considering the comments by Phil and David I think I will leave well alone and try to live with the problem until it becomes really bad then offer it for sale as "Repair or Spares" on the Forum or ebay and buy another. I didn't pay a lot for the instrument (£14 on ebay) As I said, once it is working it is OK for the rest of the day.

I have a couple of DMMs but prefer an AVO for some measurements.

Many thanks to all who offered advice on the problem.
Electrone is offline  
Closed Thread

Thread Tools



All times are GMT +1. The time now is 10:12 pm.


All information and advice on this forum is subject to the WARNING AND DISCLAIMER located at https://www.vintage-radio.net/rules.html.
Failure to heed this warning may result in death or serious injury to yourself and/or others.


Powered by vBulletin®
Copyright ©2000 - 2024, vBulletin Solutions, Inc.
Copyright ©2002 - 2023, Paul Stenning.