Old 10th Oct 2013, 8:44 pm   #1
plati1945
Retired Dormant Member
 
Join Date: Oct 2013
Location: Lowestoft, Suffolk, UK.
Posts: 4
Default Fluke 8600A Digital Multimeter Display Problem

Hi, I've recently acquire an 8600A -01. I plugged it in to try it and it just came up with dots on the display. Upon opening the case I found the batteries leaking. I replaced them with 4 new Nimh batteries but this did not improve the operation of the meter. I then starting checking and found that the +/-15V rails were not working. The 5V rail was fine. I have removed the power supply board and checked all the transistors and diodes and they all seem to be fine. Can you give me any pointers as to where I should look next? I have downloaded the manual from Fluke. The one thing that I have noticed is that the resistor which is across the 2.8uF capacitor is 100k not 1M as stated on the circuit diagram. It does look original though with no signs of additional solder etc. to suggest it has been changed. Any help will be very welcome, thanks in advance.

Last edited by AC/HL; 10th Oct 2013 at 8:53 pm. Reason: Moved to new thread
plati1945 is offline  
Old 10th Oct 2013, 11:26 pm   #2
David G4EBT
Dekatron
 
David G4EBT's Avatar
 
Join Date: Feb 2003
Location: Cottingham, East Yorkshire, UK.
Posts: 5,761
Default Re: Fluke 8600A Digital Multimeter Display Problem

I have no personal experience with Fluke meters but over the years I've seen many references to display problems and suggestions as to how to cure them. This might be of help:

http://www.eham.net/reviews/detail/3057

Other forum members with direct experience might be able to advise better on this.

Best of luck!

David
David G4EBT is offline  
Old 11th Oct 2013, 9:15 am   #3
PaulR
Dekatron
 
PaulR's Avatar
 
Join Date: Oct 2003
Location: Southport Lancashire, UK.
Posts: 3,233
Default Re: Fluke 8600A Digital Multimeter Display Problem

This thread may be of some use

https://www.vintage-radio.net/forum/...ad.php?t=84157
PaulR is offline  
Old 11th Oct 2013, 9:25 am   #4
PETERg0rsq
Heptode
 
PETERg0rsq's Avatar
 
Join Date: Jun 2009
Location: St Helens, Lancashire, UK.
Posts: 641
Default Re: Fluke 8600A Digital Multimeter Display Problem

Did you pull the resistor from the board before checking its value?
It could be the capacitor is leaking badly!
If I recall correctly it is a tantalum capacitor (I am away from home on holiday in France at moment, so cant check) and though never had a problem with one in a Fluke meter, they can prove troublesom in other devices.

As this is an inverter type PSU, the fact one side is OK eliminates the oscillator side of the PSU.

Look at the circuit and find a point after the regulator that can be lifted and check if the fault is with the PSU, or a fault elswhere in the circuit dragging the voltage down.

Batteries should be NiCad, not sure how the charging circuit will behave when confronted with NiMh. Will not cause the problems you have, but may damage the NiMh batteries. I think it uses a bulb as a current limiting device (maybe getting mixed up here with Fluke 8000 series, so check first!!!!), so not ideal for NiMh!!!


Peter
__________________
SPECIALIST.....Knows everything about nothing
EXPERT..........Knows nothing about everything

Last edited by PETERg0rsq; 11th Oct 2013 at 9:28 am. Reason: maybe getting mixed up with Fluke 8000 series!!!
PETERg0rsq is offline  
Old 11th Oct 2013, 10:00 am   #5
plati1945
Retired Dormant Member
 
Join Date: Oct 2013
Location: Lowestoft, Suffolk, UK.
Posts: 4
Default Re: Fluke 8600A Digital Multimeter Display Problem

The capacitor is not leaking, the resistor is colour coded to 100k and measures 100k. Should I change it? I have replaced NiCads with MiMh in several different pieces of equipment in the past and have never had any trouble, so I'm hoping this will be the same again. I think the problem is with the inverter circuit but I'm not sure how to check it. All the diodes, transistors and resistors seem to be OK. The transformer has no sign of any damage. It has continuity through the windings and isolation between primary and secondary.
plati1945 is offline  
Old 11th Oct 2013, 3:56 pm   #6
Leon Crampin
Octode
 
Join Date: Nov 2005
Location: Surrey, UK.
Posts: 1,870
Default Re: Fluke 8600A Digital Multimeter Display Problem

The 8600A charging circuit is quite different from the 8000A and there is no lamp to act as ballast and as a visual charging indicator. The resistor across the "ballast" capacitor in the mains primary circuit is merely a discharge to prevent the mains plug pins being live on disconnection. 100k Ohms will do fine.

If you use cells other than NiCd, you must ensure that the battery voltage (4 cells) is within TTL tolerance limits. 4 x NiCd cells give 5V +/- 0.25 which is correct.

I've given my 8600A manual to another Forum member, so I can't check finer details. These meters are well worth restoring.

Leon.
Leon Crampin is online now  
Old 11th Oct 2013, 4:23 pm   #7
plati1945
Retired Dormant Member
 
Join Date: Oct 2013
Location: Lowestoft, Suffolk, UK.
Posts: 4
Default Re: Fluke 8600A Digital Multimeter Display Problem

NiMh batteries are well within TTL tolerance, so no problem there. If you want another copy of the manual I can send you one or you can download it yourself from Fluke. If I connect a 5V DC supply to the power supply board will that be all that is needed to test it and try and find out why the inverter is not working, It looks OK to do this to me but I don't want to damage anything that may be irreplaceable due to the age of the meter.
plati1945 is offline  
Old 11th Oct 2013, 4:52 pm   #8
John_BS
Octode
 
John_BS's Avatar
 
Join Date: Dec 2008
Location: Wincanton, Somerset, UK.
Posts: 1,780
Default Re: Fluke 8600A Digital Multimeter Display Problem

If I've got the correct info, this power supply is mains in which charges a 5V battery, then a simple invertor to generate +/- higher voltages, followed by a linear regulator (U201) to give +/- 15V?
In which case, check for more than +/-15V on the output of the bridge rectifier CR203 to 206. If there's no volts there, it's the invertor (or an overload fault on the output). If there are volts, it could be U201 caput.

John
John_BS is offline  
Old 6th Nov 2013, 5:12 pm   #9
plati1945
Retired Dormant Member
 
Join Date: Oct 2013
Location: Lowestoft, Suffolk, UK.
Posts: 4
Default Re: Fluke 8600A Digital Multimeter Display Problem

I have removed the power supply board and fed it from a 5V supply. The -15V regulator was faulty so I replaced it. The board then supplied steady + 15 and -15 voltages. Upon replacing it in the meter however, the voltages were pulled down by the rest of the circuitry. After I first replaced the P/S board, I did get some flashing of the 7 segment displays together with some clicking from the relay as it switched in and out. This only lasted a few seconds and then choke L2 started to smoke! A quick switch off was needed! Any suggestion as to likely cause would be very welcome!
plati1945 is offline  
Closed Thread




All times are GMT +1. The time now is 12:32 pm.


All information and advice on this forum is subject to the WARNING AND DISCLAIMER located at https://www.vintage-radio.net/rules.html.
Failure to heed this warning may result in death or serious injury to yourself and/or others.


Powered by vBulletin®
Copyright ©2000 - 2024, vBulletin Solutions, Inc.
Copyright ©2002 - 2023, Paul Stenning.