HP334A Distortion Meter - Nulling - foam problem.
There's been a few threads concerning the problem of degrading foam eating into component leads on this forum. I have a HP334A distortion meter that wasn't nulling out the fundamental, degrading foam was the cause of one fault.
Whilst trying to trace the fault I opened a black box that contains one lamp, called a diode in the manual and a four photo resistors on a PCB as well as another lamp and single photo resistor. This forms part of the automatic rejection filter. In the lid of the box are two sunken sections which had expanded foam in them. This had degraded and eaten through the lead of one lamp. Luckily it had a tiny stub of the lead, which I was able to clean and solder a single copper wire from multistrand cable on to it. This has fixed the lamp but the meter has other issues and will need more work. However if anyone has one of these meters, it would be worth checking on the foam, and removing it. I used that black anti static foam that IC's etc come on. Andy.
__________________
Curiosity hasn't killed this cat...so far. Last edited by Diabolical Artificer; 22nd May 2017 at 9:21 am. |
22nd May 2017, 10:02 am | #2 |
Dekatron
Join Date: Apr 2012
Location: Worksop, Nottinghamshire, UK.
Posts: 5,553
|
Re: HP334A Distortion Meter - Nulling - foam problem.
Ouch that frying pan must surely be more comfortable than the fire.
|
22nd May 2017, 11:33 am | #3 |
Nonode
Join Date: Jan 2009
Location: Bristol, UK.
Posts: 2,062
|
Re: HP334A Distortion Meter - Nulling - foam problem.
Take care, it maybe a little conductive ...
dc |