Hameg HM 205-3 Ocilloscope disorted screen.
I got an oscilloscope on-line, the seller told me it was bad and the picture said so, but I am trying to repair this thing and anyone know what it could be.
Pls if anyone had this problem and they repaired this pls leave a comment on what it could be . Thanks. Last edited by AC/HL; 12th Sep 2014 at 2:43 pm. Reason: Readability |
12th Sep 2014, 3:02 pm | #2 |
Nonode
Join Date: Jun 2009
Location: Herefordshire, UK.
Posts: 2,495
|
Re: Hameg HM 205-3 Ocilloscope disorted screen.
Can please you confirm that what you are displaying in the photograph is intended to be the test calbration waveform with the oscilloscope set to a single channel mode.
|
12th Sep 2014, 6:47 pm | #3 |
Dekatron
Join Date: Feb 2003
Location: Cottingham, East Yorkshire, UK.
Posts: 5,766
|
Re: Hameg HM 205-3 Ocilloscope disorted screen.
Don't know if you have the manual, which might assist you. If not, you can download it free of charge from Hameg's website here:
http://www.hameg.com/manuals.0.html?&no_cache=1&L=0 It's very comprehensive (65 pages, 3.5MB) At the above link, in the 'PRODUCT' column, scroll down for HM203-5, and in the 'LANGUAGE' column, choose the language that is best for you out of several listed. Hope that helps. Good luck with sorting the fault - sorry I can't help further.
__________________
David. BVWS Member. G-QRP Club member 1339. |
12th Sep 2014, 9:54 pm | #4 |
Nonode
Join Date: Oct 2008
Location: Warsaw, Poland and Cambridge, UK
Posts: 2,681
|
Re: Hameg HM 205-3 Ocilloscope disorted screen.
That looks like shipping damage - I'm almost certain its been dropped and something's broken inside the CRT, distorting the electrode structure and thus the trace. The only way to fix it is to replace the CRT, I'm afraid.
Chris
__________________
What's going on in the workshop? http://martin-jones.com/ |
12th Sep 2014, 10:28 pm | #5 |
Moderator
Join Date: Mar 2012
Location: Fife, Scotland, UK.
Posts: 22,902
|
Re: Hameg HM 205-3 Ocilloscope disorted screen.
I'm afraid I think Chris is exactly right. You now need to find another of the same model with a different fault to be able to make one good one. Never send good scopes by couriers or mail. They need box within box within box packaging designed to limit peak G. Electrostatic CRTs are much more fragile than their magnetic cousins.
David
__________________
Can't afford the volcanic island yet, but the plans for my monorail and the goons' uniforms are done |
12th Sep 2014, 10:58 pm | #6 |
Dekatron
Join Date: Jun 2009
Location: Staffordshire Moorlands, UK.
Posts: 5,274
|
Re: Hameg HM 205-3 Ocilloscope disorted screen.
I'm sure there's been a similar topic on the forum before with similar symptoms? The consensus there was internal crt damage too
__________________
Kevin |
12th Sep 2014, 11:24 pm | #7 |
Octode
Join Date: May 2003
Location: Belper Derbyshire
Posts: 1,936
|
Re: Hameg HM 205-3 Ocilloscope disorted screen.
Good evening,
looks like the glass rods that hold the deflection plates in the CRT have snapped. An easy way to check is to rock and shake the scope powered up and see if the lines move around the screen. If they do the deflection plates are wobbling around on their lead in wires. The whole gun assembly would normally be held rigidly together with glass rods fused to all of the electrodes. Christopher Capener
__________________
Interests in the collection and restoration of Tefifon players and 405 line television |
12th Sep 2014, 11:42 pm | #8 |
Rest in Peace
Join Date: May 2011
Location: Preston, Lancashire UK
Posts: 955
|
Re: Hameg HM 205-3 Ocilloscope disorted screen.
I'm afraid that your crt (cathode ray tube) is damaged, and you will need a new tube from a similar scope.
Bill |
13th Sep 2014, 9:35 am | #9 |
Dekatron
Join Date: Jan 2005
Location: Oxford, UK.
Posts: 17,865
|
Re: Hameg HM 205-3 Ocilloscope disorted screen.
Yes, see here: https://www.vintage-radio.net/forum/...d.php?t=106959
|
13th Sep 2014, 10:38 am | #10 |
Retired Dormant Member
Join Date: Sep 2014
Location: Shkodër, Albania
Posts: 2
|
Re: Hameg HM 205-3 Ocilloscope disorted screen.
Well the CRT is bad so this thing will need to go for parts.
If anyone has a old analogue oscilloscope for a good price I would like to buy it because this was to be my first scope. Last edited by Mike Phelan; 14th Sep 2014 at 11:56 am. |
13th Sep 2014, 3:18 pm | #11 |
Heptode
Join Date: Mar 2009
Location: Portsmouth, Hampshire, UK.
Posts: 674
|
Re: Hameg HM 205-3 Ocilloscope disorted screen.
Just a thought but is there any chance it could be caused by extreme ripple on the power supply - enough to cause the regulator to drop out?
Caused by a shorted rectifier diode and/or dried out reservoir cap? Jim |
14th Sep 2014, 9:10 pm | #12 |
Rest in Peace
Join Date: May 2011
Location: Preston, Lancashire UK
Posts: 955
|
Re: Hameg HM 205-3 Ocilloscope disorted screen.
No, I don't think so. CRT shock damage.
|
14th Sep 2014, 10:32 pm | #13 |
Octode
Join Date: Dec 2008
Location: Wincanton, Somerset, UK.
Posts: 1,784
|
Re: Hameg HM 205-3 Ocilloscope disorted screen.
As this is a storage 'scope with more buttons than normal, it's just possible that it might not be the crt. But you'll need another 'scope to diagnose what's wrong and fix this one, assuming of course that it's not the crt. You could try removing the connections to the Y plates and see if you can get a straight line across the screen. John
|
14th Sep 2014, 10:41 pm | #14 |
Octode
Join Date: Nov 2010
Location: Ross-on-Wye, Herefordshire, UK.
Posts: 1,654
|
Re: Hameg HM 205-3 Ocilloscope disorted screen.
And while I agree that CRT internal damage is the most likely problem here, I did wonder if a strong magnetic field might be causing the trace distortion. That does seems very unlikely - since I can't think of any obvious source of such a field.
Richard |
19th Sep 2014, 10:08 am | #15 |
Octode
Join Date: May 2010
Location: Grantham, Lincolnshire, UK.
Posts: 1,177
|
Re: Hameg HM 205-3 Ocilloscope disorted screen.
I had a spectrum analyser that had similar symptoms. It had been dropped on its back and the whole electron structure in the CRT had been moved about 5mm backwards, you could see the scrape marks in the glass. A right off I am afraid, I did try banging it back but no luck.
|