Old 1st Oct 2010, 12:27 pm   #1
ronbryan
Octode
 
Join Date: Jan 2003
Location: Guildford, Surrey, UK.
Posts: 1,960
Default CT212 sig gen desiccator reactivation

This Admiralty spec AM/FM signal generator is hermetically sealed and contains a desiccator cartridge that screws in from the outside. I've recently fixed a couple of faults on it and after putting it all back together I thought I would dry out the A.P.61043 air drier cartridge and put it back into service. The equipment has been indoors for 6 months, so has been stored in dry conditions.

I removed the cartridge, taped over the hole and dried it out for an hour at 150 degrees C in our gas oven. This turned the end disc blue, so I refitted it into the generator (whilst still warm) and used it. After a few days it turned pink again, so I repeated the drying process, this time refitting the cartridge whilst both it and the equipment was still warm. Again it went pink after a few days.

The hermetic seal is ok, as you can feel pressure being released if you remove the cartridge when the sig gen is hot, so I presume the desiccator is not properly dried out, rather than more damp air is getting in. I'm now trying Igranic's suggestion of leaving the cartridge on top of the c/h boiler for a few days, but wonder what forum members think about the following questions.

Is it a bad idea to use a gas oven to dry out a desiccator cartridge (maybe the heat is too damp)?

Was an hour far too short a drying period, even though the disc went blue?

Do desiccators eventually become worn out, so that they have limited capacity and need re-charging or replacement?

Does anybody know if replacement A.P. 61043 cartridges are available?

Ron
Attached Thumbnails
Click image for larger version

Name:	Desiccator-side_web.jpg
Views:	130
Size:	67.7 KB
ID:	40953   Click image for larger version

Name:	Desiccator-end_web.jpg
Views:	129
Size:	48.0 KB
ID:	40954   Click image for larger version

Name:	CT212_web.jpg
Views:	138
Size:	115.0 KB
ID:	40955  

Last edited by ronbryan; 1st Oct 2010 at 12:41 pm. Reason: CT212 picture added
ronbryan is offline  
Old 1st Oct 2010, 12:54 pm   #2
Leon Crampin
Octode
 
Join Date: Nov 2005
Location: Surrey, UK.
Posts: 1,874
Default Re: CT212 sig gen desiccator reactivation

Ron, one problem with using a gas oven is that a considerable amount of water is produced when gas is burned. The levels of humidity inside a gas oven are high - this can have an adverse effect on pastry and other dishes which explains why electric ovens are preferred by some users.

I would guess that for your job, you would need to dry the desiccator and the whole instrument in a warm and really dry place. It might take quite a long time. Depending on domestic usage, an airing cupboard might do for the uncased instrument, but I think you might need dry heat at a higher temperature for the desiccator.

I once had a signal generator with such a unit fitted and never ever managed to get the desiccator to stay green. As it wasn't a very good generator, I didn't persevere.

Leon.
Leon Crampin is offline  
Old 1st Oct 2010, 1:37 pm   #3
medge799
Pentode
 
medge799's Avatar
 
Join Date: Feb 2009
Location: Saffron Walden, Essex, UK.
Posts: 100
Default Re: CT212 sig gen desiccator reactivation

Hi,

Dunno if this helps much, but Farnell Ireland do desiccator cartridges which might fit. I googled a bit and found a data sheet (see here http://www.farnell.com/datasheets/302226.pdf). Interstingly this implies that reactivation IS possible : '4 hours in a ventilated oven at 120degC'. I would do this when your spouse is out, though

Ta,
Mark
__________________
Mark Edgeworth
____________________________________________
BVWS Member
medge799 is offline  
Old 1st Oct 2010, 1:43 pm   #4
Hunts smoothing bomb
Octode
 
Hunts smoothing bomb's Avatar
 
Join Date: Oct 2005
Location: Wimborne, Dorset, UK.
Posts: 1,407
Default Re: CT212 sig gen desiccator reactivation

Hi Ron,

Personally I would put the sig gen and desiccator (removerd from the instrument with the hole in the panel unblocked) in a room with a dehumidifier running and seal all door and window cracks up with tape. Position the desiccator so you can view it from an outside window so you can monitor it's status, once turned blue then leave for say another two or three days for the instrument to really dry out.

I don't know if it will work but I would certainly try it.

Cheers
__________________
Lee
Hunts smoothing bomb is offline  
Old 1st Oct 2010, 1:51 pm   #5
Hunts smoothing bomb
Octode
 
Hunts smoothing bomb's Avatar
 
Join Date: Oct 2005
Location: Wimborne, Dorset, UK.
Posts: 1,407
Default Re: CT212 sig gen desiccator reactivation

One prob with just getting a new cartridge is that the original is already soaked up to capacity which also means the moisture is in the machine to the same level also, I don't think your desiccator is kaput, it is simply drawing the moisture out of the machine.

Cheers
__________________
Lee
Hunts smoothing bomb is offline  
Old 1st Oct 2010, 2:13 pm   #6
Sean Williams
Dekatron
 
Sean Williams's Avatar
 
Join Date: Feb 2005
Location: St.Ippolyts, Hitchin, Hertfordshire QRA IO91UW
Posts: 3,518
Default Re: CT212 sig gen desiccator reactivation

Put a couple of bags of silica gel inside the case for a couple of weeks, and leave the thing powered up with a vent open?
__________________
Engineers make things work and have spare bits when finished
Sean Williams is offline  
Old 1st Oct 2010, 4:18 pm   #7
Top Cap
Octode
 
Top Cap's Avatar
 
Join Date: Mar 2006
Location: Watford, Hertfordshire, UK.
Posts: 1,270
Default Re: CT212 sig gen desiccator reactivation

its usual to fill the instrument with Nitrogen (does not support moisture) before fitting the desiccator.
Les
__________________
Whether the Top Cap is Grid or Anode - touching it will give you a buzz either way!
Top Cap is offline  
Old 2nd Oct 2010, 11:45 am   #8
medge799
Pentode
 
medge799's Avatar
 
Join Date: Feb 2009
Location: Saffron Walden, Essex, UK.
Posts: 100
Default Re: CT212 sig gen desiccator reactivation

Quote:
Originally Posted by Top Cap View Post
its usual to fill the instrument with Nitrogen (does not support moisture) before fitting the desiccator.
Les
Interesting; nearly 30 years ago I used to work for Racal making military radios. When the radio came off the production line a pressure line fitting would be screwed into the desiccator hole, a low-pressure airline attached and the whole thing was dunked in a tank of water to check for leaks. When done, the desiccator cartridge would be screwed in. Is the nitrogen fill a more recent technique, I wonder?

Ta,
Mark
__________________
Mark Edgeworth
____________________________________________
BVWS Member
medge799 is offline  
Old 2nd Oct 2010, 12:23 pm   #9
Stockden
Heptode
 
Join Date: Mar 2008
Location: Forest of Dean, Gloucestershire, UK.
Posts: 648
Default Re: CT212 sig gen desiccator reactivation

I can't say that I ever encountered nitrogen filling of equipment during my military service. Certainly none of the technical workshops in the units I served with were equipped to fill equipment (Larkspur, Clansman or Ptarmigan) with any gas. Somewhere I have a copy of the Army technical manual (EMER) for the CT-212. If I can find it I'll see if it has anything to say on the matter.

As to drying out of desiccators I seem to remember that the EMER for the Wireless Set No.88 had a diagram, which detailed local manufacture of an oven for drying both equipment and desiccator. The oven essentially consisted of a large metal box containing a number (6?) of electric light bulbs as the heating element. Sadly I no longer have a copy of that document.

Hugh
Stockden is offline  
Old 2nd Oct 2010, 1:04 pm   #10
rectifier
Retired Dormant Member
 
Join Date: Sep 2010
Location: Buckingham, Bucks. UK.
Posts: 6
Default Re: CT212 sig gen desiccator reactivation

We use nitrogen fill with silica bags at work. I don't think it has to be nitrogen - any bottled gas will be dry - but oxygen would be a really bad choice if you are trying to prevent rust.
I also agree about gas ovens - they have very high humidity compared to electric ovens.

Andy.
rectifier is offline  
Old 2nd Oct 2010, 4:49 pm   #11
Stockden
Heptode
 
Join Date: Mar 2008
Location: Forest of Dean, Gloucestershire, UK.
Posts: 648
Default Re: CT212 sig gen desiccator reactivation

Well, I've located the EMER and this is what it has to say:

1. The equipment is of the sealed type and should not be removed from its case unless Oven, Drying, Telecommunications Equipment and Apparatus, Seal Testing are available.

2. (not relevant)

3. Prior to repairs and resealing the equipment should be thoroughly dried out in the oven unit of the drying equipment. When time permits, the equipment together with the case should be left in the oven for at least two hours. On removal from the oven the equipment should be sealed without delay and tests carried out as detailed in para. 4.

4. After resealing, the equipment should be tested for leakage by raising the internal air pressure to 10lb./sq. in. using the Apparatus seal testing. Immersed in water no air bubbles should be visible. In accordance with EMER TELS M631 the time constant shall be at least 100 hours with a starting pressure of 5 lb./sq. in. Dry air drawn from the appropriate outlet of the drying apparatus should be used for these tests in all cases when the ambient humidity is high. After satisfactory completion of these tests an active desiccator should be fitted.


Hugh
Stockden is offline  
Old 2nd Oct 2010, 10:08 pm   #12
ronbryan
Octode
 
Join Date: Jan 2003
Location: Guildford, Surrey, UK.
Posts: 1,960
Default Re: CT212 sig gen desiccator reactivation

Thank you everyone for your opinions, advice and EMER document research.

I'm now drying the desiccator with dry heat and in the absence of any space in the airing cupboard, have warmed up the equipment with the desiccator removed. I'll try refitting the desiccator once more before I try other ideas such as adding additional desiccator bags, because dismantling the case and reassembling is a bit of a pain.

Thanks again.

Ron
ronbryan is offline  
Old 2nd Oct 2010, 11:43 pm   #13
jimmc101
Heptode
 
jimmc101's Avatar
 
Join Date: Mar 2009
Location: Portsmouth, Hampshire, UK.
Posts: 674
Default Re: CT212 sig gen desiccator reactivation

Get a large plastic container, something like a collapsible water carrier with a tap would be ideal.
Fill with as much silica gel as you can lay your hands on but leave plenty of air in it.
Stick a bung in in the desiccator hole of the sig-gen and connect to plastic container using a length of polythene tube.

Squeeze plastic container, some dry air will be forced into sig-gen; stop squeezing, some moist air will return to plastic container.
Pause to allow this air to dry then repeat until bored rigid.

Sig-gen will now contain dry air, replace desiccator and you're done.

This is a DIY version of the method we use at work, there the pressure cycling is automatic and continues until specified dew point is reached.

Jim
jimmc101 is online now  
Old 2nd Oct 2010, 11:48 pm   #14
amp_mangler
Retired Dormant Member
 
Join Date: Aug 2010
Location: Mareeba, Queensland, Australia.
Posts: 140
Default Re: CT212 sig gen desiccator reactivation

Do you have a " paper cupboard" ?. Its the one used for keeping laser printer paper dry.
all it is, is a large resistor connected across the mains. ( I dont know the values unless I remove the element from my paper cupboard) Nitrogen is used quite a lot in modern ( and not so modern) equipment. I used work for CSIRO and all the LNB,s were filled with liquid nitrogen, but that was to keep the temperature stable, rather than keep them dry. Thats because radio telescopes work on colour temperature rather than actual "radio" signals

Joe
amp_mangler is offline  
Old 3rd Oct 2010, 12:14 am   #15
DAVEHALL
Retired Dormant Member
 
Join Date: Jan 2010
Location: Coventry, Warwickshire, UK.
Posts: 339
Default Re: CT212 sig gen desiccator reactivation

Quote:
Originally Posted by Stockden View Post
I can't say that I ever encountered nitrogen filling of equipment during my military service. Certainly none of the technical workshops in the units I served with were equipped to fill equipment (Larkspur, Clansman or Ptarmigan) with any gas. Somewhere I have a copy of the Army technical manual (EMER) for the CT-212. If I can find it I'll see if it has anything to say on the matter.
Hugh
Similarly on the 353( Clansman) production - pressure test ( from memory) was to pressurise set ,leave hose connected and dunk in a water tank .No bubbles -remove and seal .And seem to remember it was only just dry air.
DAVEHALL is offline  
Closed Thread




All times are GMT +1. The time now is 10:32 pm.


All information and advice on this forum is subject to the WARNING AND DISCLAIMER located at https://www.vintage-radio.net/rules.html.
Failure to heed this warning may result in death or serious injury to yourself and/or others.


Powered by vBulletin®
Copyright ©2000 - 2024, vBulletin Solutions, Inc.
Copyright ©2002 - 2023, Paul Stenning.