If I get a chance later, I'll be having a bit of lab time, I'll probe it again and take a photo and show you.

The other thing I want to do for the sake of it is the last crystal calibrator check. From the manual on page 38 under Crystal Calibrator checks, 3C. It says I need to connect an audio oscillator to pin one of the RF box tag strip. Now, there are a few tag strips in the RF box, the monitor tag strip has two strips numbered 1 to 10 on the drawings and another strip inside the box not numbered. It says to connect via a capacitor, value to be decided/suggested.

But first, find the pin one!
__________________
Avometer, vintage Fluke and Marconi collector. Also interested in vintage Yaesu and KW.
Sinewave is offline  
Old 24th Dec 2019, 10:30 pm   #142
Sinewave
Octode
 
Sinewave's Avatar
 
Join Date: May 2016
Location: Oxfordshire/Bucks borders, UK.
Posts: 1,604
Default Re: Marconi TF144H/4S

Here's the 2Mhz Xtal waveform.
Attached Thumbnails
Click image for larger version

Name:	TF144H4S 2mhzXtal.jpg
Views:	65
Size:	117.8 KB
ID:	195825  
__________________
Avometer, vintage Fluke and Marconi collector. Also interested in vintage Yaesu and KW.
Sinewave is offline  
Old 24th Dec 2019, 11:07 pm   #143
G0HZU_JMR
Dekatron
 
Join Date: Sep 2010
Location: Cheltenham, Gloucestershire, UK.
Posts: 3,077
Default Re: Marconi TF144H/4S

Thanks. At first glance that looks like the waveform I'd expect to see at the cathode at pin 8. At the cathode I'd expect to see a kind of wonky sine wave but with a flat shape to the negative peaks. This would correspond to the anode current falling to zero for that part of the RF cycle.

At the grid on pin 7 I'd hope to see a fairly good sine wave at several volts rms assuming that the probe doesn't spoil things here. Do you see a sine wave at pin 7?

I see you have a Tek TDS2012/2 scope. I've got one of those here and the counter on it is very good and stable. This is the dedicated frequency counter readout in the lower right corner of the display rather than the menu counter further up on your display. Much better than the vague and jittery software based counter found in many digital scopes!
__________________
Regards, Jeremy G0HZU

Last edited by G0HZU_JMR; 24th Dec 2019 at 11:15 pm.
G0HZU_JMR is offline  
Old 24th Dec 2019, 11:40 pm   #144
Sinewave
Octode
 
Sinewave's Avatar
 
Join Date: May 2016
Location: Oxfordshire/Bucks borders, UK.
Posts: 1,604
Default Re: Marconi TF144H/4S

Its the TDS2014, the counter on it is very useful, very stable. Very convenient scope and bench space saving. Managed to pick it up for a bargain price when a company I were working for made us all redundant and needed to get the building empty so we all managed to get items at silly prices. I wouldn't have had it otherwise as they aren't cheap, I've got a few cathode ray scopes which although basic, are rather adequate, though I do find the features on this scope very useful indeed, such as measurement, cursor, the trigger menu items and the compactness, portability and space saving is a real bonus.

I've just scoped pin 7 and I do get a nice clean sinewave in the order of 2.15VRMS.
__________________
Avometer, vintage Fluke and Marconi collector. Also interested in vintage Yaesu and KW.
Sinewave is offline  
Old 24th Dec 2019, 11:51 pm   #145
G0HZU_JMR
Dekatron
 
Join Date: Sep 2010
Location: Cheltenham, Gloucestershire, UK.
Posts: 3,077
Default Re: Marconi TF144H/4S

I bought my 2 channel TDS2012 not long after they were first released. This was in the early 2000s and I got it at the (then) bargain price of about £550. These scopes don't get good reviews on Eevblog but I like them a lot for certain tasks. They are small and easy to use and they are silent. If I'm writing any MCU code I always use this scope because I can work in total silence in the room. The only niggle with mine is that it only has 2 channels which often isn't enough for MCU work. the TDS2014 is nicer because it has 4 channels.
__________________
Regards, Jeremy G0HZU
G0HZU_JMR is offline  
Old 25th Dec 2019, 12:01 am   #146
Sinewave
Octode
 
Sinewave's Avatar
 
Join Date: May 2016
Location: Oxfordshire/Bucks borders, UK.
Posts: 1,604
Default Re: Marconi TF144H/4S

They do now have 2 channel scopes at reasonable prices. I think these ones were in the order of about 2k a pop at the time our company bought them.

Once I'm happy with this 'pin 1' to connect and audio oscillator up to, it's nearing completion of the checks. I have some other signal generators, rather than audio oscillators, I'm hoping they'll do.
__________________
Avometer, vintage Fluke and Marconi collector. Also interested in vintage Yaesu and KW.
Sinewave is offline  
Old 25th Dec 2019, 12:14 am   #147
G0HZU_JMR
Dekatron
 
Join Date: Sep 2010
Location: Cheltenham, Gloucestershire, UK.
Posts: 3,077
Default Re: Marconi TF144H/4S

They might be referring to the connection '1' in the little square box at the top of the crystal calibrator circuit at FILT 1. This will have a 22k connection (R216) to 250V DC so some care would be needed here along with a high voltage coupling cap. This node in the circuit corresponds to the filtered audio output from the heterodyne mixer at the top of FILT 1. So it would seem logical that this would be the place to inject a test audio signal... but maybe not the safest!

If this is the right connection then I'd have expected a safety warning in the manual because there could be 100-250V DC via R216 here. It might give you a jolt and it might also zap any (modern low voltage) test gear you connect to it even with a DC blocking cap in the way.

They also ask for the range control to be set to a 'between 2 ranges' setting. Will this partial rotation of the range switch put the switch wafer in an open circuit position to isolate the mixer stage and prevent it loading the test point?
__________________
Regards, Jeremy G0HZU

Last edited by G0HZU_JMR; 25th Dec 2019 at 12:26 am.
G0HZU_JMR is offline  
Old 25th Dec 2019, 12:24 am   #148
Sinewave
Octode
 
Sinewave's Avatar
 
Join Date: May 2016
Location: Oxfordshire/Bucks borders, UK.
Posts: 1,604
Default Re: Marconi TF144H/4S

On page 33, I were thinking it were one of those tag strips, which are inside the RF box.
__________________
Avometer, vintage Fluke and Marconi collector. Also interested in vintage Yaesu and KW.
Sinewave is offline  
Old 25th Dec 2019, 12:35 am   #149
G0HZU_JMR
Dekatron
 
Join Date: Sep 2010
Location: Cheltenham, Gloucestershire, UK.
Posts: 3,077
Default Re: Marconi TF144H/4S

I think it might be the tag strip shown next to figure 4.8 on page 39. This looks to be the same location I described in post #147.

I'm also uncertain what "select CRYSTAL CHECK RANGE Control between two ranges" actually means. Do they mean that the control is deliberately fudged to rest halfway between two range positions?
__________________
Regards, Jeremy G0HZU

Last edited by G0HZU_JMR; 25th Dec 2019 at 12:41 am.
G0HZU_JMR is offline  
Old 25th Dec 2019, 12:46 am   #150
Sinewave
Octode
 
Sinewave's Avatar
 
Join Date: May 2016
Location: Oxfordshire/Bucks borders, UK.
Posts: 1,604
Default Re: Marconi TF144H/4S

Yes I think so. By the looks of what the test requires, it's to make sure that any 'range' isn't connected, so that any tests injecting voltage is diverted away from other circuits.
__________________
Avometer, vintage Fluke and Marconi collector. Also interested in vintage Yaesu and KW.
Sinewave is offline  
Old 26th Dec 2019, 1:51 pm   #151
Sinewave
Octode
 
Sinewave's Avatar
 
Join Date: May 2016
Location: Oxfordshire/Bucks borders, UK.
Posts: 1,604
Default Re: Marconi TF144H/4S

Boxing day isn't for cold turkey, it's for fettling and pottering in the 'lab'.

Right, so having located pin 1, it was a tag strip outside of the RF box, just under the HV regulator valves.

So, with Crystal Check selected and Range set to A, I measure at Pin 1, 112V DC. With range between two ranges, 86V DC.

So, that voltage will need to be considered if I'm going to connect a signal generator to it.
__________________
Avometer, vintage Fluke and Marconi collector. Also interested in vintage Yaesu and KW.
Sinewave is offline  
Old 26th Dec 2019, 2:08 pm   #152
Sinewave
Octode
 
Sinewave's Avatar
 
Join Date: May 2016
Location: Oxfordshire/Bucks borders, UK.
Posts: 1,604
Default Re: Marconi TF144H/4S

A 100V 1nF poly cap will probably be fine for coupling to the circuit, it'll block the DC but allow a small AC signal through.
__________________
Avometer, vintage Fluke and Marconi collector. Also interested in vintage Yaesu and KW.
Sinewave is offline  
Old 26th Dec 2019, 10:00 pm   #153
Sinewave
Octode
 
Sinewave's Avatar
 
Join Date: May 2016
Location: Oxfordshire/Bucks borders, UK.
Posts: 1,604
Default Re: Marconi TF144H/4S

So, it sort of works. The ALC is certainly doing something, but the levels I can't be sure of at the moment as the presence of any RF may be why Marconi are specifying that this measurement needs to be done with the valve volt meter.

My valve volt meter (Marconi TF1041C) is out of calibration, it does need some work doing to it, but I can roughly gauge that it is giving different results in these areas, but can't be certain of the actual reading, so at some point I will need to sort that out.
__________________
Avometer, vintage Fluke and Marconi collector. Also interested in vintage Yaesu and KW.
Sinewave is offline  
Old 28th Dec 2019, 9:22 pm   #154
Sinewave
Octode
 
Sinewave's Avatar
 
Join Date: May 2016
Location: Oxfordshire/Bucks borders, UK.
Posts: 1,604
Default Re: Marconi TF144H/4S

So, there's still a few things to do, however the unit is working fine really, as reported earlier, but I still wanted to go through the whole set up.

So regarding voltages, the valve voltmeter is being specified to measure some of these voltages I'm using a modern DVM for, which may be getting upset with present RF.

For the record with my DVM I'm measuring at the phones socket up to 38V DC with my signal generator connected (pin 1 tag strip) at 1khz 20v pk-pk (manual suggests these are DC voltages, however the rectified probe on a valve volt meter is for AC measurements). The max should be 20v and if not then the manual suggests checking V204, MR206 and C208. All appear to be fine. With carrier off this can be 65-70v.

I don't actually have an audio oscillator, just regular generators, but the sinewave appears pretty clean.

So, to try and be sure what I shall do is come back to this with a working valve voltmeter (hopefully) and see if this can be nailed down. Then hopefully I'll be somewhat satisfied with the Crystal Calibrator section of this set up procedure/performance checks.
__________________
Avometer, vintage Fluke and Marconi collector. Also interested in vintage Yaesu and KW.
Sinewave is offline  
Old 2nd Aug 2020, 5:44 pm   #155
Sinewave
Octode
 
Sinewave's Avatar
 
Join Date: May 2016
Location: Oxfordshire/Bucks borders, UK.
Posts: 1,604
Default Re: Marconi TF144H/4S

So, since sorting out a valve volt meter, I returned to a few puzzling areas on the TF144H/4S.

There were some questions over measured voltages and whether or not RF being present was having any affect on the digital meters readings. So I got hold of a TF2604 valve volt meter (working), expecting things to be different.

For now, I continued with the 'Performance Checks' section, rather than at this time going back to the voltages measured from the drawings, which weren't a part of the 'Performance Checks' section.

Going back to

Performance checks: Crystal Calibrator.

Previous measured voltages were,

Range E: 29.35V
Range A: 19.44V

They were measured with a DVM, I obtained the same results with a VTVM.

The performance checks suggested 2.5 to 16V each.

If incorrect, check or adjust crystal is the only suggestion from the manual.

So, drawings have incorrect voltages recorded? Or is something else happening which I've missed.

There are some other odd things to add in a moment too...
__________________
Avometer, vintage Fluke and Marconi collector. Also interested in vintage Yaesu and KW.
Sinewave is offline  
Old 15th Aug 2020, 6:59 pm   #156
Sinewave
Octode
 
Sinewave's Avatar
 
Join Date: May 2016
Location: Oxfordshire/Bucks borders, UK.
Posts: 1,604
Default Re: Marconi TF144H/4S

So, I'm wondering if previously replaced resistors have had anything to do with any of these differing voltage measurements. As there were some values which were different according to colour code than what the drawing said they were.

The other one here is the ALC voltage measured at the phones out jack, with an oscillator had previously measured 38VDC. Now I've picked this up with the valve meter, I'm still not seeing anything different. In fact this time it's a little higher at 50V and then dropping to 20V when the connected oscillator is set to 20V

The connected oscillator is 20v pk-pk, though the manual doesn't say if that should have been pk-pk.

Some confusing results, but otherwise the generator appears to be working. I've scoped the set up to veiw the modulation and that appears ok, there are checks to make from ranges C to L, though I can't set modulation at all on range C.

Any suggestions?
__________________
Avometer, vintage Fluke and Marconi collector. Also interested in vintage Yaesu and KW.
Sinewave is offline  
Old 16th Aug 2020, 2:48 pm   #157
Sinewave
Octode
 
Sinewave's Avatar
 
Join Date: May 2016
Location: Oxfordshire/Bucks borders, UK.
Posts: 1,604
Default Re: Marconi TF144H/4S

Additionally, despite a few odd voltages here and there, as it works well as a generator, could I be asking too much for everything to be perfect on a 1960s generator?
__________________
Avometer, vintage Fluke and Marconi collector. Also interested in vintage Yaesu and KW.
Sinewave is offline  
Old 16th Aug 2020, 7:58 pm   #158
Radio Wrangler
Moderator
 
Radio Wrangler's Avatar
 
Join Date: Mar 2012
Location: Fife, Scotland, UK.
Posts: 22,902
Default Re: Marconi TF144H/4S

No reason why it shouldn't be possible to bring it up to full spec, if you want to. Provided that damnable thermocouple is intact.

Whether you need the full performance and whether you see the effort as worthwhile remain the questions you must answer.

I once told a man, who while surveying the mangled remains of his pride and joy, had just said it must be a write off and couldn't possibly be repaired ... "Sir, it was once an iron ore deposit, so it must be possible to rebuild it. The issue is one of whether it's worth doing."

They're actually a good generator, but I did change from one of them to a TF2008.

David
__________________
Can't afford the volcanic island yet, but the plans for my monorail and the goons' uniforms are done
Radio Wrangler is offline  
Old 1st Sep 2020, 8:14 pm   #159
Sinewave
Octode
 
Sinewave's Avatar
 
Join Date: May 2016
Location: Oxfordshire/Bucks borders, UK.
Posts: 1,604
Default Re: Marconi TF144H/4S

Quote:
Originally Posted by Radio Wrangler View Post
No reason why it shouldn't be possible to bring it up to full spec, if you want to. Provided that damnable thermocouple is intact.

Whether you need the full performance and whether you see the effort as worthwhile remain the questions you must answer.

I once told a man, who while surveying the mangled remains of his pride and joy, had just said it must be a write off and couldn't possibly be repaired ... "Sir, it was once an iron ore deposit, so it must be possible to rebuild it. The issue is one of whether it's worth doing."

They're actually a good generator, but I did change from one of them to a TF2008.

David
Gerernally speaking, it appears to be working well. Maybe I'm getting too hung up on some of these voltages and wondering why? The out of tolerance resistors have been replaced, that leaves capacitors and valves and I have tried other valves and no difference was nothed in the results.

As some have already said pages ago, they believe from either their knowledge/previous experience, that some of the voltages in the servicing manual to be incorrect. Otherwise, the generator is working with the only aparrent fault being modulation not working on a few of the ranges, not a big deal.

Output sinewave is clean, apart from a bit more poking around to find why the modulation isn't switching in range C, (I think A - C), then there's not much more I can do. I will however pick another one up at the rallies next year if I see one and see if there are any differences.
__________________
Avometer, vintage Fluke and Marconi collector. Also interested in vintage Yaesu and KW.
Sinewave is offline  
Old 2nd Sep 2020, 11:57 am   #160
Andrew Sinclair
Pentode
 
Join Date: Sep 2007
Location: Fareham, Hampshire, UK.
Posts: 105
Default Re: Marconi TF144H/4S

I used one of these generators as my main shack RF signal generator from 1990 to 2011. The oscillator valve is quite critical. I found some samples of 5763 (or QV03-12) work much better than others. I thought either Mullard or Brimar worked better but can't remember which... The ALC voltages were quite different from one sample to another, even though they looked similar on the AVO CT160 valve tester. I stopped using this generator due to the sleeving used in the coil turrets deteriorating and going conductive. It was too big a job to dismantle the turret and change it.
I replaced many resistors and a few capacitors in mine and it ran reliably most of the time I owned it.
Andrew Sinclair is offline  
Closed Thread




All times are GMT +1. The time now is 7:51 am.


All information and advice on this forum is subject to the WARNING AND DISCLAIMER located at https://www.vintage-radio.net/rules.html.
Failure to heed this warning may result in death or serious injury to yourself and/or others.


Powered by vBulletin®
Copyright ©2000 - 2024, vBulletin Solutions, Inc.
Copyright ©2002 - 2023, Paul Stenning.