Old 29th Feb 2020, 11:55 am   #1
RevGeoff
Default Farnell L30-2
Hello,
I have acquired the above, and the current range does not show an indication, although the voltage is variable, the current limit works and it delivers power to a load.
I suspect VT101 - 2N5294, or VT3 - 2N3053. Comments before I open it up?
RevGeoff is offline
Old 29th Feb 2020, 1:03 pm   #2
short wave
Default Re: Farnell L30-2
Hi RevGeof,
if the voltage is variable and is present at the O/P the control circuit is ok.
I would suspect the switch(if single meter type) and associated current circuit resistors(across the main current sensing resistor)
short wave is offline
Old 29th Feb 2020, 3:57 pm   #3
RevGeoff
Default
Hi.
I did wonder about the switch. Is the sense resistor R206?
I also forgot to state that the current meter setting pot has no effect.
RevGeoff is offline
Old 1st Mar 2020, 11:35 am   #4
short wave
Yes ��
Old 1st Mar 2020, 11:58 am   #5
Wendymott
Wendymott's Avatar
If you have output.. the current sense resistor will be ok..Either the switch or any series resistor FROM the current sense to the meter.
Should get out more.
Regards
Wendy G8BZY
Wendymott is offline
Old 5th Mar 2020, 11:19 pm   #6
RevGeoff
Success!
I replaced the bridge rectifier, as C1 was getting hot, suggesting it was passing AC. I replaced C1 in case it was damaged. I also replaced the switch. VT101 was OK, but VT3 was intermittent, also replaced. This seems to have cured the problem. Now to soak test it.
Thanks guys!
RevGeoff is offline