Telequipment D65 scope faulty
Telequipment D65 scope
Got this of my local freecycle group it’s in need of repair Never repaired a scope so need some guidance and a scan of the service manual please The fault is on the screen I only have an unfocused 1 cm wide vertical line that runs from top to bottom I can move it left to right with the position pot but that’s all Cheers Vince |
17th Jun 2013, 8:30 am | #2 |
Octode
Join Date: Feb 2007
Location: Ely, Cambridgeshire, UK.
Posts: 1,976
|
Re: Telequipment D65 scope faulty
have managed to get the D65/D66 service manual a scan of fair qualty
just need a pointer in the right direction |
17th Jun 2013, 8:35 am | #3 |
Dekatron
Join Date: Mar 2008
Location: Worthing, West Sussex, UK
Posts: 5,185
|
Re: Telequipment D65 scope faulty
Hi Vince, I have a pdf copy of the manual, if you would like it please PM me your email address.
Mark |
17th Jun 2013, 11:21 am | #4 |
Rest in Peace
Join Date: May 2011
Location: Preston, Lancashire UK
Posts: 955
|
Re: Telequipment D65 scope faulty
Sounds like the timebase is not running. First job on any instrument such as this is to check all the power supply rails against the manual and get that right first. If you don't check all the psu outputs for correct voltage and ripple, you will be chasing your tail. Be careful when dealing with the eht section.
Bill |
17th Jun 2013, 11:34 am | #5 |
Octode
Join Date: Feb 2007
Location: Ely, Cambridgeshire, UK.
Posts: 1,976
|
Re: Telequipment D65 scope faulty
cheers will start there
and report back |
17th Jun 2013, 11:24 pm | #6 |
Nonode
Join Date: Dec 2012
Location: Kirk Michael, Isle of Man
Posts: 2,350
|
Re: Telequipment D65 scope faulty
I don't remember exactly, but I think the mains incoming is exposed at the fuse. Possibly a bigger danger than the EHT? I usually put some sort of plastic insulator over the fuseholder. TQ also did a mod whereby a plastic lid was retrofitted over the voltage selector plug/socket. If you are idiot enough, you could poke a bit of wire, or even a wet pencil lead in there and get a tickle, hence that mod.
Les. |
18th Jun 2013, 7:23 am | #7 |
Octode
Join Date: Feb 2007
Location: Ely, Cambridgeshire, UK.
Posts: 1,976
|
Re: Telequipment D65 scope faulty
Tthanks Lles good tip.
Wwill look at doing the same. |
19th Jun 2013, 8:01 am | #8 |
Octode
Join Date: Feb 2007
Location: Ely, Cambridgeshire, UK.
Posts: 1,976
|
Re: Telequipment D65 scope faulty
OKok got back to this it has already had a new mains lead fitted and the fuse holder insulated.
Aall power rails measure OKok and there is very little ripple on them. I did notice when Ii power off the display becomes a dot that falls from the top of the display to the centre and fades away. The dot size changes depending on how the focus control is set. |
19th Jun 2013, 8:29 am | #9 |
Rest in Peace
Join Date: May 2011
Location: Preston, Lancashire UK
Posts: 955
|
Re: Telequipment D65 scope faulty
That sounds like the Y amp is unbalanced. Basically its a wideband amplifier with a push-pull output operating on the Y plates of the crt. You need to check the output voltages on the output to the plates, you will probably find that you have a severe imbalance. If so check the output transistors. Voltages in this area must be as per the manual.Check also that all transistors are present and if socketed, give them a twiddle to ensure good contact.
Also, if the spot doesn't move when the timebase control is advanced, then there is a fault in the timebase too. Bill |
19th Jun 2013, 9:31 am | #10 |
Octode
Join Date: Feb 2007
Location: Ely, Cambridgeshire, UK.
Posts: 1,976
|
Re: Telequipment D65 scope faulty
will have a look cheers
|