Default AVO VCM 163 Meters

Hi all,

Just doing a few checks on my VCM 163, after removing the meters from the VCM and checking current required for FSD, the Ia meter indicates FSD at 21 uA and the mA/V meter indicates FSD is at 35 uA - am I missing something obvious here? These are not doing what I expect, shouldn't FSD be at 50 uA for both these meters - any ideas?

Meter resistance values are correct, the low uA required for FSd would also explain why the meters are off scale when selected to check AC, MA/V.

I have replaced the set AC / current meter with a 2.4k resistor as I don't want to damage it during circuit testing, the current measured is around 45 uA which is I believe required for the meter pointer to reach the set AC / calibration position

Any help appreciated.