hello Dave,
1.) check -150V
first of all you have to check the voltages from the PSU, always !
And always at first the -150 volts bec. this is used for reference to provide all other voltages in that PSU. If there is another then -150V , there is a pot to adjust it, and if this will not help write here again. Adjust the -150V very exactly, <1%

Tell us if you have a running scope there, if yes check the ripple on the -150V also,
if you have a high voltage test head for the multimeter or VTVM?
if you have a valves tester?

2.) check all other voltages from the PSU

+100V
+225V
+350V
+500V

if possible give us a screen picture and of coarse the results of your test.

greetings from germany
Martin

! take care, the PSU is very powerful, life dangerous voltages !