What year is it? If it's an early one:

The 10k ohm range relies on a resistance of around 180 ohm.. a photo would be a help as i expected this 180 ohm to be a winding but are you saying it's a component resistor.?

The 1000 ohm range uses a winding of around 15 ohm. Although it's failure to function could also be a switching problem. There's an extra pair of contacts that have to close on this particular range, as i recall.

There is (i think) a third winding on the resistance board, which is about 1 ohm (quite a thick one)

The high range (100,000 ohm) can be a nuisance as the Q potentiometer has inherent problems especially in old age- the Q shaft must have continuity with the cleverly shaped brass orifice in the facia. Also there is a little arrow shaped precious metal prod in the end of the Q shaft, and it's not unknown for this to be slightly loose, which can affect all 3 ranges. I posted some images re. this issue so if you search under 'Q shaft' or something like that you will find them in the avometer survey thread, or test equipment thread.