Old 12th Apr 2021, 7:59 pm   #162
SATOLEX
Triode
 
Join Date: Feb 2021
Location: Leiden, Zuid-Holland, The Netherlands
Posts: 26
Default Re: Spectrum Analyser TR4172

Hi all, I have fixed the broken inductor ground connections in the 206 MHz band-pass filter and have trimmed the amplifier-filter combination. This solved my 14 dB gain loss issue in the RF part, see picture . I have a gain adjustment margin of about +1.5 dB (cal. screw) and this is ok for the moment. I managed to do the auto calibration and for what I have seen so far is that it does not freeze anymore. I have trimmed the 206 MHz BPF based on the service manual instructions on page 11-65. According my paper manual version the maximum gain ripple has to be smaller than 0.3 dB, a symmetric response with a maximum level. The picture in the manual suggests that it has to be flat in a certain bandwidth but this not specified. I’m not sure what the right response has to be, I assume a narrow bandwidth, see my picture. Jeremy do you want to perform one more measurement for me, please? You only have to connect the tracking to input-1 and press some key’s. I’m curious how your 206 MHz BPF response looks like.
Best regards,
Lex
Attached Thumbnails
Click image for larger version

Name:	TR4172_cal-20dBm fixed.jpg
Views:	93
Size:	91.2 KB
ID:	231626   Click image for larger version

Name:	BPF 206 MHz adjustment.jpg
Views:	95
Size:	81.6 KB
ID:	231627  
SATOLEX is offline