Default Re: Marconi Valve Voltmeter TF1041C - repair thread

Ok so I have arrived at a possible PSU problem.

The HT+ line is stated as +230v in the manual, but the regulator can only muster +220v despite having +276v on its anode.
Likewise the HT- is stated as -120, but it's only producing -112V

Adjusting the slider RV12 from 16v to 3v has virtually no effect on the HT voltage, whereas the manual says this should be set to produce +230v HT+.

The voltage between R59 and C7 is also 220v (anode of V6b)

See diagram below.
The resistors are within tolerance and C7 was replaced.

Does this look like a soft 12BH7 valve to you guys?


I'm not sure of this is causing the dc offset problem, but it's a all I have at the moment.

Thanks