HV probe definitely!

The 2.5kv range on the AVO takes too much current and loads down scope HT circuitry.

There are some very high impedances in the CRT bias networks. All the 175s had been replaced by 180s before my time at HP, but if they follow the usual practice, there may be dual2.7kv supplies, one for grid one for cathode so that a lower voltage amplifier can blank the trace when the timebase is not triggered.