Re: I've injured my Keithley 175 DMM
Removing the -V supply to U103 settles the instrument down, the display reads 00.00 on any voltage range but it does respond to the range pushbuttons so I think the µP is working.
U109 may be the culprit but I've got no further than clipping the +V leg to that which made no difference to the -7V. I have ordered a couple of TLC271 chips to replace U103 - not only is that the regulator for the -7V but it's also in close proximity to the probe connections on the front panel.