Re: Hameg HM307-3 Cleaning and Repair
X final amplifier measurements (the acceptable range is 84.1V-90.1V):
C of TR113- 112V
C of TR114- 96V
Both are unbalanced, like the Y amp.
R181 and R182 measure spot on out of circuit. The voltages on TR111 and TR112 are correct.
CRT section investigation:
85V missing on E of T354. R368 was O/C. Replacing it restored E voltage.
-8.9V was found on pin 4 (-590V--680V) of the CRT. The focus potentiometer has a partly broken inner ring. It is still partly attached. I have glued it.
All of the other voltages are present. The are either a bit too high or a bit too low.
I'll see if the pot is salvageable in the morning.
|