It just says to allow the instrument to stabilize and perform the test at an ambient temperature of 23 C.

On the 20 volt DC range it says in the chart that a 19.00 volt test voltage is acceptable to have a display reading between 18.97 to 19.03 volts. I'm a bit surprised that this is acceptable, but then it is an early generation meter. I'm sure it's capable of better.

I need to really do some proper checks of all the ranges to see if it's a general inaccuracy over the whole set.

I'm very pleased with the meter and it'll get some proper use on the bench once I'm happy with its accuracy.