Default Re: Hameg HM208 Oscilloscope Timebase fault

Hi

Further Pics added:

Pic (f) Timebase set to 0.5us/cm

Pic (g) Timebase set to 0.2us/cm

Pic (h) The Add/Chop and Dual buttons depressed and an apparently normal display is shown even at low timebase speeds.

Pic (i) A 1kHz Calibration signal applied to CH1 input, Timebase set to 2us/cm

Pic (j) A 1MHz Calibration signal applied to CH1 input, Timebase set to 0.2us/cm.


Note how far off accuracy the trace is particularly in Pic (i) This signal is occupying approx 5.3cm so T = 10.6us, therefore f = 94.3kHz. The signal is meant to be 1kHz. I checked this calibration signal on my Scopex scope and all's well there.
The inaccuracy seems to improve as the timebase speed is increased. Its as if there's non linearity in the trace. I get the impression that the trace is being compressed at low timebase speeds.

Any further help with this will be much appreciated. Thanks.

Regards
Symon.
Attached Thumbnails
Click image for larger version

Name:	(f) 0.5us per cm.jpg
Views:	68
Size:	40.8 KB
ID:	153673   Click image for larger version

Name:	(g) 0.2us per cm.jpg
Views:	81
Size:	43.5 KB
ID:	153674   Click image for larger version

Name:	(h) Add Chop and Dual buttons depressed.jpg
Views:	59
Size:	45.6 KB
ID:	153675   Click image for larger version

Name:	(i) 1kHz Calibration signal 2us per cm.jpg
Views:	144
Size:	47.6 KB
ID:	153676   Click image for larger version

Name:	(j) 1MHz Calibration signal 0.2us per cm.jpg
Views:	85
Size:	51.2 KB
ID:	153677  

Philips210 is offline