Old 22nd Oct 2014, 11:12 am   #1
mimmo2010
Retired Dormant Member
 
Join Date: Jul 2014
Location: Rome, Italy.
Posts: 23
Default Power supply HP182T

hi, I have a problem with the power supply Hp182T, after about 10 minutes of operation, the correct value of +100 volts starts giving values ​​ranging from 88 volts to 97 volts and not 'more stabilized, may depend on what ??
regards
mimmo2010 is offline  
Old 22nd Oct 2014, 11:44 am   #2
threeseven
Octode
 
threeseven's Avatar
 
Join Date: May 2010
Location: Worcestershire, UK.
Posts: 1,880
Default Re: Power supply HP182T

Have you checked the output voltage with an external meter? What type of meter does the power supply have?
__________________
BVWS member
threeseven is offline  
Old 22nd Oct 2014, 12:11 pm   #3
Herald1360
Dekatron
 
Herald1360's Avatar
 
Join Date: Feb 2007
Location: Leominster, Herefordshire, UK.
Posts: 16,535
Default Re: Power supply HP182T

HP182T is a scope mainframe.

Manuals seem to be available from the usual suspects.....
__________________
....__________
....|____||__|__\_____
.=.| _---\__|__|_---_|.
.........O..Chris....O
Herald1360 is offline  
Old 22nd Oct 2014, 2:46 pm   #4
Humber888
Heptode
 
Humber888's Avatar
 
Join Date: Jul 2012
Location: Bromley, London, UK.
Posts: 702
Default Re: Power supply HP182T

All the LV power supplies are referenced to the +100V one. The +100V supply gets its reference from a temperature compensated zener. So, are the other supplies varying as well? Where are you measuring the +100V rail? The best place is at the +100V test point on the mainframe regulator board. This is easily accessed by removing the locking cover on the bottom rear of the 'scope. (Don't remove the cover while the 'scope is on.) Also, is your measurement taken with the plug-ins inserted? If so, switch off, unplug them, switch on and then see if the 100V rail continues to vary.

Mike
Humber888 is offline  
Old 23rd Oct 2014, 8:15 am   #5
mimmo2010
Retired Dormant Member
 
Join Date: Jul 2014
Location: Rome, Italy.
Posts: 23
Default Re: Power supply HP182T

Hello Mike, I have already changed the electrolytic capacitors in the output, a resistor A1A2R9 33,2k be replaced with BC337 transistor 2n3704 A1A2Q2 fault because it worked fine a couple of days with stable voltages, then began to make the default. I think it is not good that the BC337 !!
I found the horizontal amplifier and two resistors routes A5R38 A5R55 from 22,1k.Le test I'm doing a vacuum extract the power from Hp182t.
I tried plugging a lamp as load to 60watt although there are 90 volt output will not turn on.
volts have been taken in the points test
regards Thanks
Mimmo
mimmo2010 is offline  
Old 28th Oct 2014, 11:19 am   #6
mimmo2010
Retired Dormant Member
 
Join Date: Jul 2014
Location: Rome, Italy.
Posts: 23
Default Re: Power supply HP182T

Sorry for the delay, everything is ok, the problem is solved I replaced the BC337 with a 2N3704 and the voltage is back to normal.
Thanks to all
mimmo2010 is offline  
Closed Thread

Thread Tools



All times are GMT +1. The time now is 1:25 pm.


All information and advice on this forum is subject to the WARNING AND DISCLAIMER located at https://www.vintage-radio.net/rules.html.
Failure to heed this warning may result in death or serious injury to yourself and/or others.


Powered by vBulletin®
Copyright ©2000 - 2024, vBulletin Solutions, Inc.
Copyright ©2002 - 2023, Paul Stenning.